Look back autoscore settings

Postby Sleep_research » Wed Feb 24, 2010 4:34 am

I am new to Compumetics and learning as I go. I have reviewed the books and followed the directions available there but am having a problem with the autoscore settings in the look back screen.

What I have done: Tools - Options - and under autoscore I have turned it OFF. We are still experiencing autoscoring for plm's, arousals and even spindles. When you open the same study outside of look back we are not having this problem - only in the lookback screen.

Does anyone have any insight as to what can be done to turn this off in lookback? Is there a different setting that needs to be changed?

Thanks!
Sleep_research
New Member
 
Posts: 1
Joined: Thu Oct 29, 2009 1:21 am

Postby bduce » Fri Jul 16, 2010 1:22 am

By stating that you are turned off the autoanalysis algorithms in Tools/Options means that you are trying to fix it through ProFusion. That is your review/analysis program. You need to go to PSG Config and turn it off in the study configuration setting. The PSG Config program dictates what PSG Online will do during the recording, including the autoanalysis during recording.

Hope that helps.
Brett